A dispensing balance passes at 200 g but fails near a 2 g working load. Can it remain in use for 2 g API dispensing?
Direct answer
Not without evidence that the low-load requirement is met or suitable restrictions and controls are implemented.
Detailed explanation
- direct answer
Not without evidence that the low-load requirement is met or suitable restrictions and controls are implemented.
- scientific and process reasoning
Separate high-load calibration success from low-load capability, evaluate precision and bias near dispensing quantity, protect affected material and document alternate balance selection or validated control.
- Evidence to review
The decision should be supported by relevant approved acceptance criteria, calibration method, as-found/as-left records where applicable, reference-standard certificates, uncertainty information and a documented GMP impact assessment. Case-specific evidence: Separate high-load calibration success from low-load capability, evaluate precision and bias near dispensing quantity, protect affected material and document alternate balance selection or validated control.
- illustrative application
ILLUSTRATIVE — At a hypothetical GMP site, a quality reviewer must decide whether a near-limit reading can be accepted. The site evaluates objective evidence, intended use and product risk rather than treating a calibration sticker as definitive proof.
- common interpretation error
High-capacity compliance proves minimum-weight suitability.
- Regulatory context
US CGMP 21 CFR 211.160(b)(4) requires calibration at suitable intervals under a written program specifying accuracy/precision limits and remedial action; 21 CFR 211.68(a) addresses calibration or checks of relevant automatic/mechanical/electronic equipment. NIST guidance distinguishes metrological traceability from fitness for purpose, and ISO/IEC 17025:2017 concerns calibration-laboratory competence. Exact controls and decision rules depend on jurisdiction, measurement application and approved site procedures; these references do not mandate a universal annual interval or universal uncertainty ratio.
Source references — not yet verified
- Source register — 21 CFR 211.160(b)(4) — Laboratory controls and calibration
Source supplied with the draft; version and applicability not verified.
Source register link — unverified applicability - Source register — 21 CFR 211.68(a) — Automatic, mechanical, and electronic equipment
Source supplied with the draft; version and applicability not verified.
Source register link — unverified applicability - Source register — JCGM 200:2012 — International Vocabulary of Metrology (VIM), 3rd ed.
Source supplied with the draft; version and applicability not verified.
Source URL requires editorial verification.
- Source register — NIST — Measurement Uncertainty
Source supplied with the draft; version and applicability not verified.
Source URL requires editorial verification.
Review and applicability
Scientific review incomplete. No scientific reviewer, approval date or regulatory endorsement is claimed.
Educational information. Apply the relevant jurisdiction, product context and approved site quality system.